Dear Günter,

I think I have a dirty fix for this.

In drasi and ucesb, use the branch 'quickflush'.

--

Please test on some PC not running a DAQ or data transport:

Make sure the 'tmux' program is installed.

Make three terminals.  The first needs to be wide and tall:

# First: this sets up a simulated DAQ with 30 Hz triggers.
# This runs inside tmux.  To exit, press 'C-b d'.

cd drasi/
make empty -j 32
scripts/runsim.sh --trigrate=30 --slaves=0 --timesort --trans

# Second: takes data from the end of the simulated DAQ chain, and sends 
# using ext_data interface.

cd ucesb/
make empty -j 32
empty/empty --trans=localhost --ntuple=SERVER,dummy

# Third: read data using ext_data interface:

cd ucesb/
ext_test/ext_reader_h101 localhost

# The third should now show a more or less continous stream of updates:
# (not be 'chunked')

       2880 (d         1):  2
       2881 (d         1):  2
       2882 (d         1):  2

# The simulated DAQ can be killed with:

cd drasi/
scripts/runsim.sh --kill

--

To implement this more permanently, I would need to understand again how 
they pace the data flow...  In order to not build more unnecessary options 
on top of old.

Cheers,
Håkan



On Mon, 22 Jul 2024, Weber, Guenter Dr. wrote:

> 
> Dear friends,
> 
> 
> we are currently in a situation where we would like to use the SIS3316
> modules like a scope, i. e. once a trigger happens we would like the data
> out of the DAQ immediately. Would it be possible to modify DRASI in a way
> that it flushes out the data after each trigger? (Or, maybe easier, to set
> the minimum flush rate to something like 10 Hz?)
> 
> 
> Thank you very much!
> 
> 
> 
> 
> Best greetings
> 
> Günter
> 
> 
> 
> 
>